[Rtk-users] FDK for planar ct

Robert Calliess robert.calliess at gmx.de
Wed Nov 15 14:26:11 EST 2017


Hello,

Indeed the trajectory is very small. 

Thank you very much for your help so far. 

 

Kind regards,

Robert

 

Von: simon.rit at gmail.com [mailto:simon.rit at gmail.com] Im Auftrag von Simon Rit
Gesendet: Mittwoch, 15. November 2017 17:29
An: Robert Calliess
Cc: Cyril Mory; rtk-users at public.kitware.com
Betreff: Re: [Rtk-users] FDK for planar ct

 

Hi,

FYI, I have opened the images. As far as I can guess from the sequence, the circular trajectory of the source is very small wrt to the object size. I don't know what I would do from such a trajectory. I would start by checking the work in tomosynthesis to find some inspiration.

Good luck!

Simon

 

On Wed, Nov 1, 2017 at 10:22 AM, Robert Calliess <robert.calliess at gmx.de> wrote:

Hello,

i’ve uploaded the projection images for the first trajectory.

https://www.dropbox.com/s/rpcyemuvfngd4rw/QFN1.7z?dl=0

 

Inside the folder you can find the projection images. Furthermore you will find

two textfiles, positions.txt and calibration.txt. The positions.txt contains the physical

position of the detector center. Position id 35 is the orthogonal view. All other projection ids

are positions on a circular path around the center position 35. 

The calibration.txt contains the fod, odd and the physical detector size.

All units given are in micrometers.

 

Kind regards,

Robert

 

 

Von: Cyril Mory [mailto:cyril.mory at creatis.insa-lyon.fr] 
Gesendet: Mittwoch, 25. Oktober 2017 09:34
An: Robert Calließ
Cc: rtk-users at public.kitware.com


Betreff: Re: [Rtk-users] FDK for planar ct

 

Dear Robert,

Could you send two datasets of projections, one for each case ? We would have a look at them and it would help us understand your trajectories. The drawings you sent do not seem to be sufficient to remove all ambiguities.

Best regards,

Cyril

 

On 25/10/2017 09:03, "Robert Calließ" wrote:

Hello,

the object is moving on a circular path. There are arrows between the different detector positions showing the

moving direction. The source is static. 

 

Kind regards,

Robert

  

Gesendet: Dienstag, 24. Oktober 2017 um 16:47 Uhr
Von: "Simon Rit"  <mailto:simon.rit at creatis.insa-lyon.fr> <simon.rit at creatis.insa-lyon.fr>
An: "Robert Calließ"  <mailto:Robert.Calliess at gmx.de> <Robert.Calliess at gmx.de>
Cc:  <mailto:rtk-users at public.kitware.com> "rtk-users at public.kitware.com"  <mailto:rtk-users at public.kitware.com> <rtk-users at public.kitware.com>
Betreff: Re: Re: [Rtk-users] FDK for planar ct

Hi,

I see one drawing only, not two. And the object does not seem to be moving on your drawing, is it? If not and the source is also static (as it seem), this is equivalent to one large projection.

Simon

  

On Tue, Oct 24, 2017 at 3:58 PM, "Robert Calließ" <Robert.Calliess at gmx.de> wrote: 

Hello,

I suppose there are still misunderstandings with respect to the trajectory.

Attached you can find the two difference trajectories. I also had a closer look to

the off centered fdk ( the paper you suggested). But I don't think it is in my case.

The iso ray passes object center and detector center at each view. Off centered fdk

has a different preweighting scheme.

 

You said that the RTK ramp filter is along the u axis (orthogonal to rotaion axis). For planar_ct_1 trajectory that

should fit. As you can see at the picture, the object is moving on a circular path but not rotating around the

center point (red cross in the image).

 

 

Kind regards,

Robert C.

  

Gesendet: Donnerstag, 12. Oktober 2017 um 07:14 Uhr
Von: "Simon Rit" <simon.rit at creatis.insa-lyon.fr>
An: "Robert Calliess" <robert.calliess at gmx.de>
Cc: "rtk-users at public.kitware.com" <rtk-users at public.kitware.com> 


Betreff: Re: [Rtk-users] FDK for planar ct

Hello,

No. The filter should be orthogonal to the rotation axis.  The RTK ramp filter is along the u axis of the projection.

Trajectory 2: if you take photos by rotating the cameras, they are photographies of the same point-of-view. This is what I meant.

Cheers,

Simon

  

On Wed, Oct 11, 2017 at 8:58 PM, Robert Calliess <robert.calliess at gmx.de> wrote: 

Hello,

thanks for the link to the paper but I dont have access to it. Aside from how the trajectory is interpreted within in RTK. My actual question was

if any of those two trajectories would need another reconstruction filter than the FDK Filter. From my point of understanding a specific rotation around

the object is necessary for fbp/fdk (like c-arm bow, standard circular cone-beam trajectory).  That’s why I asked If the first trajectory needs some other reconstruction 

filter because the object itself doesn’t rotate around itself. It actually gets translated on a circular path. So I was more expecting a “yes” or “no” to the fdk filter 

or a hint to another filter (except iterative reconstructions) I should use for these trajectories.

 

To trajectory 2: I think the projections are different. The object rotates and each projection shows a different view.

 

 

Kind regards,

Robert C.

 

Von: simon.rit at gmail.com [mailto:simon.rit at gmail.com] Im Auftrag von Simon Rit
Gesendet: Dienstag, 10. Oktober 2017 20:29
An: Robert Calliess


Cc: Cyril Mory; rtk-users at public.kitware.com
Betreff: Re: [Rtk-users] FDK for planar ct

 

 

Hi,

Let me try to clarify what I mean by "source trajectory wrt the object." In tomography, you need to determine the source trajectory in the object coordinate system, we don't really care about the source trajectory in the room coordinate system. For example, rotating the source on a circular trajectory or rotating the object makes no difference for the reconstruction algorithm. That's why we call diagnostic scanners "helical scanners".

So for trajectory 1, it seems that the source trajectory (again, wrt to the object) is a circle but the object is offset. This is somewhat similar to https://doi.org/10.1109/TNS.2006.880977 except that the detector is not tilted so FDK would be the only FBP algorithm I could think of. But the situation is really not good, data are missing and iterative reconstruction should give better results.

Trajectory 2: what I said in my previous email is true, it's useless I believe, all projections are similar up to a 2D transform of the projection.

Simon

 

On Tue, Oct 10, 2017 at 8:07 PM, Robert Calliess <robert.calliess at gmx.de> wrote:

Hello,

I try to clarify the both trajectories.

 

Trajectory 1:

No, i dont move the source on two circles. The xray source is fixed. Only the object and the detector moves. Both move on a circular path so that the iso-ray

always passes through the pcb centre and the detector centre. There is one orthogonal view and the others are the ones moving on the circular path. 

(Object is not rotating around its own axis).

 

Trajectory 2:

Yes, the xray source lies in the rotation axis and only the object rotates around its z-axis. Detector and xray source are fixed and the detector is tilted.

It’s almost like this trajectory here https://www.ikeda-shoponline.com/engctsoft/wp-content/uploads/2016/06/Oblique-View-CT1.jpg

except that the xray source lies on the rotation axis.

 

I hope this helps to understand the trajectories I have to deal with.

 

Kind regards, 

Robert

 

 

Von: simon.rit at gmail.com [mailto:simon.rit at gmail.com] Im Auftrag von Simon Rit
Gesendet: Dienstag, 10. Oktober 2017 19:06
An: Robert Calließ
Cc: Cyril Mory; rtk-users at public.kitware.com


Betreff: Re: [Rtk-users] FDK for planar ct

 

Hi,

It's still not clear to me but what is helpful is to think in terms of source trajectory wrt the object.

Trajectory 1: if I understand, you move the source on two circles plus one point. I don't know of a FBP algorithm to reconstruct this, but there might be one. I would consider iterative reconstruction first.

Trajectory 2: your trajectory is a point, the source does not move with respect ot the object since it lies on the rotation axis. So each projection contains exactly the same information up to a simple 2D projection deformation. So it's hopeless to reconstruct from one projection only.

To create the correct geometry, I would suggest using the function AddProjection <https://github.com/SimonRit/RTK/blob/master/code/rtkThreeDCircularProjectionGeometry.h#L92>  for which you provide the source and detector positions plus the 3D coordinates of the two axes of the coordinate system of the projection.

I hope this helps

Simon

 

On Tue, Oct 10, 2017 at 5:43 PM, "Robert Calließ" <Robert.Calliess at gmx.de> wrote:

Hello,

thank you for the fast reply.

To answer your questions first.

In this case the abbrevation pcb stands for printed circuit board.

Next point is the trajectory we are currently handling with.

Please see the attached image "trajectory.png". There are two schematics showing the side view and top view for trajectory type 1

and a side-view for trajectory type 2.

 

For type 1:

The xray source is fixed. The pcb is clamped within a transport, so the pcb and the detector are moveable with in the xy plane.

As you can see at the image, the pcb moves along a circular path but the pcb itself is not rotating. And let's assume that the iso ray

always passes through the centre of the pcb and the centre of the detector.

 

For type 2:

The xray source is fixed and the detector is tilted. The pcb lies centred in the middle of a table. So that the pcb rotates around its centre

around the z-axis.

 

 

I hope this makes clear what trajectory i'm dealing with. Thank you.

 

Kind regards,

Robert C.

 

 

 

 

 

 

Gesendet: Dienstag, 10. Oktober 2017 um 15:31 Uhr
Von: "Cyril Mory" <cyril.mory at creatis.insa-lyon.fr>
An: "Robert Calliess" <robert.calliess at gmx.de>, rtk-users at public.kitware.com
Betreff: Re: [Rtk-users] FDK for planar ct

Dear Robert,

Your description of the trajectory is very obscure to me. Maybe you have a very unusual X-ray system. Could you make the following points clear :

- what is a PCB ?

- what is fixed/moving in your system (we need this information for the object, the source and the detector), and what kind of trajectories have the moving parts ?

- can you re-draw your sketch with just 2 or 3 positions (ideally, on similar but separate drawings), each one with the object, the source and the detector ?

If you do that, we should have a clear understanding of how your acquisition goes, and be able to give you appropriate advice.

Best regards,

Cyril

  

On 10/10/2017 15:02, Robert Calliess wrote:

Hello rtk users,

I have question to the RTK FDK Filter. As far as I understand from to the fourier slice theorem the object to be reconstructed needs a circular trajectory and needs to rotate its own centre.

Please have a look at the attached sketch. With this planar trajectory (Object, a PCB, is moved on a circle trajectpry  “in-plane”, PCB itself is not rotating) do I need

a special filtering if I want to use FDK for planar CT with respect to the sketched trajectory ? I tried a circular in-plane trajectory where the PCB is centred and rotates

around its centre point. And with 100 projections I get good results. But with the trajectory I described (sketch, attached image) the results are not so good.

Because of the row-wise ramp filter It looks like there is a directional dependency. My assumption is, and with respect to fourier slice theorem, that the missing object

rotation (rotation around itself) causes there directional effects.

 

So my questions to the experts are. Do I need to apply a special filtering before backprojecting with FDK or is it just the wrong

algorithm for this kind of trajectory ?

 

kind regards,

Robert C.

      

_______________________________________________
Rtk-users mailing list
Rtk-users at public.kitware.com
http://public.kitware.com/mailman/listinfo/rtk-users


_______________________________________________
Rtk-users mailing list
Rtk-users at public.kitware.com
http://public.kitware.com/mailman/listinfo/rtk-users

 

 





_______________________________________________
Rtk-users mailing list
Rtk-users at public.kitware.com
http://public.kitware.com/mailman/listinfo/rtk-users

 


_______________________________________________
Rtk-users mailing list
Rtk-users at public.kitware.com
http://public.kitware.com/mailman/listinfo/rtk-users

 

-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://public.kitware.com/pipermail/rtk-users/attachments/20171115/d2e41d9b/attachment-0001.html>


More information about the Rtk-users mailing list