[Rtk-users] Reg: CBCT 1k reconstruction error
Cyril Mory
cyril.mory at creatis.insa-lyon.fr
Wed Nov 2 08:27:18 EDT 2016
I see two errors:
Remove this line
writer->SetNumberOfStreamDivisions(4);
since only the streaming filter, not the write filter, needs to know
about the number of divisions.
And change the filename to
"D:\\Output.mhd"
It will write both the .mhd and the .raw files.
I hope it helps,
Cyril
On 11/02/2016 11:57 AM, Zahid Hasan Ansari wrote:
> Dear Cyril Mory,
>
> I have modified my file writer code and used StreamingImageFilter. Now its not giving any error message but the final output is 0 byte for 1024 volume. It working fine for 512 volume.
>
> Please check my modified file writer code given below.
>
> // Streaming depending on streaming capability of writer
> typedef itk::StreamingImageFilter<OutPutImageType, OutPutImageType> StreamerType;
> StreamerType::Pointer streamerBP = StreamerType::New();
> streamerBP->SetInput(feldkamp->GetOutput());
> streamerBP->SetNumberOfStreamDivisions(4);
>
> //create a writer and write reconstructed file
> itk::ImageFileWriter<OutPutImageType>::Pointer writer;
> writer = itk::ImageFileWriter<OutPutImageType>::New();
> writer->SetFileName("D:\\Output.raw");
> writer->SetImageIO(io);
> writer->SetInput(streamerBP->GetOutput());
> writer->SetNumberOfStreamDivisions(4);
>
>
> Please let me know if any clarification is required.
>
>
> Thanks & Regards
>
> Zahid Hasan Ansari
> Senior Design Engineer
> Mobile No. +91-9738379729
>
> Panacea Medical Technologies Pvt. Ltd.
> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area Phase 1, Whitefield, Bangalore - 560 066, Karnataka, INDIA
> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial Area, Malur - 563130. Kolar District. INDIA.
> Tel : +91 80 4242 8700 / 2845 4785 | Fax : + 91 42428710 | Url : http://www.panaceamedical.com
> Bangalore - India.
>
>
>
>
> ----- Original Message -----
> From: Cyril Mory [mailto:cyril.mory at creatis.insa-lyon.fr]
> To: Zahid Hasan Ansari [mailto:zahidhasan.a at panaceamedical.com], Simon Rit [mailto:simon.rit at creatis.insa-lyon.fr]
> Cc: rtk-users at public.kitware.com, saimahesh.m at panaceamedical.com
> Subject: Re: [Rtk-users] Reg: CBCT 1k reconstruction error
>
>
>> Dear Zahid,
>>
>> The --divisions is an option of the command-line application rtkfdk. It
>> is not directly an option of the FDKConeBeamReconstructionFilter :
>> rather, it is passed to a streaming filter at the end of the pipeline.
>> Look for the following bit of code in rtkfdk.cxx :
>>
>> // Streaming depending on streaming capability of writer
>> typedef itk::StreamingImageFilter<CPUOutputImageType,
>> CPUOutputImageType> StreamerType;
>> StreamerType::Pointer streamerBP = StreamerType::New();
>> streamerBP->SetInput( pfeldkamp );
>> streamerBP->SetNumberOfStreamDivisions( args_info.divisions_arg );
>>
>> You can find information on how a streaming filter works on this page:
>>
>> https://itk.org/Doxygen/html/classitk_1_1StreamingImageFilter.html
>>
>> I hope it helps,
>>
>> Cyril
>>
>> On 11/02/2016 09:01 AM, Zahid Hasan Ansari wrote:
>>> Dear Cyril Mory,
>>>
>>> I have tried but I am not getting where to give
>> --division 4 in my code.
>>>
>>> Please find my code given below and please let me know
>> where I need to change in my code to make it work.
>>>
>>> int _tmain(int argc, _TCHAR* argv[])
>>> {
>>> #pragma region "Variable declaration"
>>> const double PI = 3.14159265358979323846;
>>> float *angles;
>>> int nProj = 349;
>>> typedef unsigned short pixelType;
>>> typedef float OutpixelType;
>>> const int dimension = 3;
>>> typedef itk::Image<pixelType, dimension> imageType;
>>> typedef itk::ImageRegionConstIterator<imageType > ImageIteratorType;
>>> imageType::Pointer Projections = imageType::New();
>>>
>>> #pragma endregion "Variable declaration"
>>>
>>> Projections =
>> ReadRawImages("D:\\CBCT\\CIRS_1204020_25022016153227\\Proj_%d.raw", nProj);
>>> imageType::SpacingType pSpacing;
>>> pSpacing[0] = 0.2960;
>>> pSpacing[1] = 0.2960;
>>> pSpacing[2] = 0.2960;
>>> Projections->SetSpacing(pSpacing);
>>>
>>> imageType::PointType pOrigin;
>>> pOrigin[0] = -212.972;
>>> pOrigin[1] = -212.972;
>>> pOrigin[2] = -212.972; //-158.50;
>>> Projections->SetOrigin(pOrigin);
>>>
>>> Projections->Update();
>>>
>>> //Read angles.
>>> char *angles_file = "D:/CBCT/CIRS_1204020_25022016153227/angle.txt";
>>> angles = new float[nProj];
>>> angles = readAngles(angles_file, angles);
>>>
>>> #pragma region"Geometry"
>>> // Geometry object
>>> typedef rtk::ThreeDCircularProjectionGeometry GeometryType;
>>> GeometryType::Pointer geometry = GeometryType::New();
>>> for (unsigned int noProj = 0; noProj < nProj; noProj++)
>>> {
>>> geometry->AddProjection(1000.0, 1500.0, angles[noProj], 15.0, 10, 0, 0);
>> // 136 half fan //2.07 Kidwai //
>>> }
>>>
>>> typedef rtk::ThreeDCircularProjectionGeometryXMLFileWriter
>> GeometryWriterType;
>>> GeometryWriterType::Pointer geometryWriter = GeometryWriterType::New();
>>> geometryWriter->SetFilename("D:\\geo.xml");
>>> geometryWriter->SetObject(geometry);
>>> geometryWriter->WriteFile();
>>> geometry->Update();
>>> #pragma endregion "Geometry"
>>>
>>> //Define output image type
>>>
>>> #ifdef USE_CUDA
>>> typedef itk::CudaImage< OutpixelType, dimension > OutPutImageType;
>>> typedef rtk::ParkerShortScanImageFilter< OutPutImageType > PSSFType;
>>> typedef rtk::CudaFDKConeBeamReconstructionFilter FDKType;
>>> #else
>>> typedef itk::Image<OutpixelType, dimension> OutPutImageType;
>>> typedef rtk::ParkerShortScanImageFilter<OutPutImageType > PSSFType;
>>> typedef rtk::FDKConeBeamReconstructionFilter<OutPutImageType> FDKType;
>>> #endif
>>>
>>> //ScatterCorrection
>>> typedef rtk::BoellaardScatterCorrectionImageFilter<imageType, imageType >
>> BoellaardScatterCorrectionImageFilterType;
>>> BoellaardScatterCorrectionImageFilterType::Pointer ScatterCorrection =
>> BoellaardScatterCorrectionImageFilterType::New();
>>> ScatterCorrection->SetInput(Projections);
>>>
>>> //VarianObiRawImageFilter
>>> typedef rtk::VarianObiRawImageFilter< imageType, OutPutImageType >
>> RawImageFilterType;
>>> RawImageFilterType::Pointer AttenuationFilter =
>> RawImageFilterType::New();
>>> AttenuationFilter->SetInput(ScatterCorrection->GetOutput());
>>>
>>> //Create the output image
>>> typedef rtk::ConstantImageSource< OutPutImageType >
>> ConstantImageSourceType;
>>> ConstantImageSourceType::PointType origin_p;
>>> ConstantImageSourceType::SizeType size_p;
>>> ConstantImageSourceType::SpacingType spacing_p;
>>> ConstantImageSourceType::Pointer projectionsSource =
>> ConstantImageSourceType::New();
>>> origin_p[0] = -127.5;
>>> origin_p[1] = -127.5;
>>> origin_p[2] = -127.5;
>>> size_p[0] = 512;
>>> size_p[1] = 512;
>>> size_p[2] = 512;
>>> spacing_p[0] = (origin_p[0] / size_p[0]) * 2 * (-1);
>>> spacing_p[1] = (origin_p[1] / size_p[1]) * 2 * (-1);
>>> spacing_p[2] = (origin_p[2] / size_p[2]) * 2 * (-1);
>>>
>>> projectionsSource->SetOrigin(origin_p);
>>> projectionsSource->SetSpacing(spacing_p);
>>> projectionsSource->SetSize(size_p);
>>> projectionsSource->SetConstant(0);
>>>
>>> // Short scan image filter
>>> PSSFType::Pointer pssf = PSSFType::New();
>>> pssf->SetInput(AttenuationFilter->GetOutput());
>>> pssf->SetGeometry(geometry);
>>> pssf->InPlaceOff();
>>> std::cout << "short scan image filter success" << std::endl;
>>>
>>> FDKType::Pointer feldkamp = FDKType::New();
>>> feldkamp->SetInput(0, projectionsSource->GetOutput());
>>> feldkamp->SetInput(1, pssf->GetOutput());
>>> feldkamp->SetGeometry(geometry);
>>> feldkamp->GetRampFilter()->SetTruncationCorrection(1.0);
>>> feldkamp->GetRampFilter()->SetHannCutFrequency(5.0);
>>> feldkamp->GetRampFilter()->SetHannCutFrequencyY(5.0);
>>> feldkamp->GetRampFilter()->SetCosineCutFrequency(5.0);
>>> feldkamp->GetRampFilter()->SetHammingFrequency(5.0);
>>> feldkamp->GetRampFilter()->SetRamLakCutFrequency(5.0);
>>> feldkamp->SetGPUEnabled(1);
>>> feldkamp->SetNumberOfThreads(10000);
>>>
>>> #pragma region "Write volume"
>>>
>>> //Create a raw io for writing
>>>
>>> itk::RawImageIO<float, dimension>::Pointer io;
>>> io = itk::RawImageIO<float, dimension>::New();
>>>
>>> for (unsigned int i = 0; i < dimension; i++)
>>> {
>>> io->SetDimensions(i, size_p[i]);
>>> io->SetSpacing(i, spacing_p[i]);
>>> io->SetOrigin(i, origin_p[i]);
>>> }
>>> io->SetHeaderSize(0);
>>> io->SetByteOrderToLittleEndian();
>>> io->SetPixelType(itk::ImageIOBase::SCALAR);
>>> io->SetNumberOfComponents(1);
>>> io->SetNumberOfDimensions(3);
>>>
>>> //create a writer and write reconstructed file
>>> itk::ImageFileWriter<OutPutImageType>::Pointer writer;
>>> writer = itk::ImageFileWriter<OutPutImageType>::New();
>>> writer->SetFileName("D:\\Output.raw");
>>> writer->SetImageIO(io);
>>> writer->SetInput(feldkamp->GetOutput());
>>>
>>> try
>>> {
>>> writer->Update();
>>> }
>>> catch (itk::ExceptionObject & excp)
>>> {
>>> std::cerr << "Error while writing the image " << std::endl;
>>> std::cerr << excp << std::endl;
>>> getchar();
>>> }
>>> #pragma endregion "WriteVolume"
>>> delete[] angles;
>>> return 0;
>>> }
>>>
>>>
>>>
>>> Thanks & Regards
>>>
>>> Zahid Hasan Ansari
>>> Senior Design Engineer
>>> Mobile No. +91-9738379729
>>>
>>> Panacea Medical Technologies Pvt. Ltd.
>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area Phase 1,
>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial Area,
>> Malur - 563130. Kolar District. INDIA.
>>> Tel : +91 80 4242 8700 / 2845 4785 | Fax : + 91 42428710 | Url :
>> http://www.panaceamedical.com
>>> Bangalore - India.
>>>
>>>
>>>
>>>
>>> ----- Original Message -----
>>> From: Cyril Mory [mailto:cyril.mory at creatis.insa-lyon.fr]
>>> To: Zahid Hasan Ansari [mailto:zahidhasan.a at panaceamedical.com], Simon Rit
>> [mailto:simon.rit at creatis.insa-lyon.fr]
>>> Cc: rtk-users at public.kitware.com, saimahesh.m at panaceamedical.com
>>> Subject: Re: [Rtk-users] Reg: CBCT 1k reconstruction error
>>>
>>>
>>>> Dear Zahid,
>>>>
>>>> I was able to perform a reconstruction from your header and geometry
>>>> file, using the following command lines:
>>>>
>>>> rtkprojectshepploganphantom -o simulatedprojections.mha -g geo.xml
>>>> --phantomscale "128,128,128" --like Output.mhd
>>>>
>>>> rtkfdk -p . -r simulatedprojections.mha -g geo.xml -o fdk.mha --hardware
>>>> cuda --dimension 1024 --spacing 0.25 --lowmem --divisions 4
>>>>
>>>> The first computes projections through a Shepp & Logan phantom, with the
>>>> same size, spacing, origin, etc... as your projections, using your
>>>> geometry file geo.xml.
>>>>
>>>> The second line performs the FDK reconstruction. I had to use both the
>>>> "--lowmem" and the "--divisions 4" since I have little GPU memory (3GB,
>>>> on a Geforce GTX 780). "--lowmem" loads the projections into memory by
>>>> subsets of 16. "--divisions 4" cuts the reconstructed volume into 4
>>>> parts, reconstructs them one by one, then assembles the results. Can you
>>>> run the same commands and let us know whether you still encounter the
>>>> crash you mentioned ? If it works, you can use your own projection data
>>>> in the second command line instead of "simulatedprojections.mha".
>>>>
>>>> Best,
>>>>
>>>> Cyril
>>>>
>>>> On 10/25/2016 01:13 PM, Zahid Hasan Ansari wrote:
>>>>
>>>>> Dear Sir,
>>>>>
>>>>> Please find the Header file in the attachment.
>>>>>
>>>>>
>>>>> Thanks & Regards
>>>>>
>>>>> Zahid Hasan Ansari
>>>>>
>>>>>
>>>>> ----- Original Message -----
>>>>> From: Cyril Mory [mailto:cyril.mory at creatis.insa-lyon.fr]
>>>>> To: Zahid Hasan Ansari [mailto:zahidhasan.a at panaceamedical.com], Simon
>> Rit
>>>> [mailto:simon.rit at creatis.insa-lyon.fr]
>>>>> Cc: rtk-users at public.kitware.com, saimahesh.m at panaceamedical.com
>>>>> Subject: Re: [Rtk-users] Reg: CBCT 1k reconstruction error
>>>>>
>>>>>
>>>>>> Dear Zahid,
>>>>>>
>>>>>> We do not need the projections file, at least not for a first stage of
>>>>>> error tracking. We only need the header.
>>>>>> Try the following command line:
>>>>>>
>>>>>> rtkprojections -p . -r yourProjectionsFileName -o proj.mhd
>>>>>>
>>>>>> It should write a "proj.mhd" and a "proj.raw". The "proj.mhd" file is
>>>>>> what we need, and it is a very light text file. The "proj.raw" contains
>>>>>> the pixel values, but at the moment we do not need them. We will create
>>>>>> our own proj.raw file, filled with zeros, which should be enough to
>>>>>> track down the error you encounter.
>>>>>>
>>>>>> Looking forward to receiving your file,
>>>>>> Cyril
>>>>>>
>>>>>> On 10/25/2016 09:22 AM, Zahid Hasan Ansari wrote:
>>>>>>> Dear Sir,
>>>>>>>
>>>>>>> The single raw file of the projections is 1.4 GB and we
>> are
>>>> not
>>>>>> able to send this big file to you. Can you please suggest other
>>>> alternatives
>>>>>> for this?
>>>>>>> Or can you provide us the PC configuration to solve the
>>>> issue?
>>>>>>> Let me know if any other clarification is required.
>>>>>>>
>>>>>>>
>>>>>>> Thanks & Regards
>>>>>>>
>>>>>>> Zahid Hasan Ansari
>>>>>>>
>>>>>>>
>>>>>>>
>>>>>>> ----- Original Message -----
>>>>>>> From: Cyril Mory [mailto:cyril.mory at creatis.insa-lyon.fr]
>>>>>>> To: Zahid Hasan Ansari [mailto:zahidhasan.a at panaceamedical.com], Simon
>>>> Rit
>>>>>> [mailto:simon.rit at creatis.insa-lyon.fr]
>>>>>>> Cc: rtk-users at public.kitware.com, saimahesh.m at panaceamedical.com
>>>>>>> Subject: Re: [Rtk-users] Reg: CBCT 1k reconstruction error
>>>>>>>
>>>>>>>
>>>>>>>> Hello Zahid,
>>>>>>>>
>>>>>>>> We will need the header of your projections file, too (It is best if
>>>> you
>>>>>>>> have all your projections as a single .mhd and a single .raw file, so
>>>> it
>>>>>>>> should be 3-D image, and you send only the .mhd file).
>>>>>>>>
>>>>>>>> Regards,
>>>>>>>> Cyril
>>>>>>>>
>>>>>>>> On 10/19/2016 03:18 PM, Zahid Hasan Ansari wrote:
>>>>>>>>> Dear Cyril Mory,
>>>>>>>>>
>>>>>>>>>
>>>>>>>>> I have only used RTK version 1.3.0. but in the
>> error
>>>>>>>> message it is showing RTK version 1.2.0.
>>>>>>>>> Please find the attachment of the geometry file
>> of
>>>> our
>>>>>>>> projections.
>>>>>>>>> Thanks & Regards
>>>>>>>>>
>>>>>>>>> Zahid Hasan Ansari
>>>>>>>>>
>>>>>>>>>
>>>>>>>>>
>>>>>>>>> ----- Original Message -----
>>>>>>>>> From: Cyril Mory [mailto:cyril.mory at creatis.insa-lyon.fr]
>>>>>>>>> To: Zahid Hasan Ansari [mailto:zahidhasan.a at panaceamedical.com],
>> Simon
>>>>>> Rit
>>>>>>>> [mailto:simon.rit at creatis.insa-lyon.fr]
>>>>>>>>> Cc: rtk-users at public.kitware.com, saimahesh.m at panaceamedical.com
>>>>>>>>> Subject: Re: [Rtk-users] Reg: CBCT 1k reconstruction error
>>>>>>>>>
>>>>>>>>>
>>>>>>>>>> Dear Zahid,
>>>>>>>>>>
>>>>>>>>>> Without some more information, it's unlikely that we find the
>> source
>>>> of
>>>>>>>>>> the problem. Here are a few things you can do to help us (and
>>>> therefore
>>>>>>>>>> yourself):
>>>>>>>>>> - try version 1.3.0 of RTK: there have been considerable changes in
>>>> the
>>>>>>>>>> Cuda forward and back projection filters since v1.2.0, so your
>>>> problem
>>>>>>>>>> might disappear just by upgrading to the new version
>>>>>>>>>> - create a small example that reproduces your problem. You can, for
>>>>>>>>>> example, simulate a geometry, simulate projections of a shepp logan
>>>>>>>>>> phantom, and reconstruct from these projections (take a look at
>>>>>>>>>> http://wiki.openrtk.org/index.php/RTK/Scripts/FDK if you need an
>>>>>>>>>> example). And send us the script
>>>>>>>>>> - OR send us your geometry file and the header of your projections
>>>> file
>>>>>>>>>> (no need to send the projection data itself, we'll create a
>>>> zero-filled
>>>>>>>>>> stack of projections), and the command line that crashes
>>>>>>>>>>
>>>>>>>>>> Best regards,
>>>>>>>>>> Cyril
>>>>>>>>>>
>>>>>>>>>> On 10/17/2016 02:27 PM, Zahid Hasan Ansari wrote:
>>>>>>>>>>> Dear Simon Rit,
>>>>>>>>>>>
>>>>>>>>>>> Please find the screen shot in the attachment.
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>> Thanks & Regards
>>>>>>>>>>>
>>>>>>>>>>> Zahid Hasan Ansari
>>>>>>>>>>> Senior Design Engineer
>>>>>>>>>>> Mobile No. +91-9738379729
>>>>>>>>>>>
>>>>>>>>>>> Panacea Medical Technologies Pvt. Ltd.
>>>>>>>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area
>>>>>> Phase
>>>>>>>> 1,
>>>>>>>>>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>>>>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial
>>>>>> Area,
>>>>>>>>>> Malur - 563130. Kolar District. INDIA.
>>>>>>>>>>> Tel : +91 80 4242 8700 / 2845 4785 | Fax : + 91 42428710 | Url :
>>>>>>>>>> http://www.panaceamedical.com
>>>>>>>>>>> Bangalore - India.
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>> ----- Original Message -----
>>>>>>>>>>> From: Simon Rit [mailto:simon.rit at creatis.insa-lyon.fr]
>>>>>>>>>>> To: Zahid Hasan Ansari [mailto:zahidhasan.a at panaceamedical.com]
>>>>>>>>>>> Cc: rtk-users at public.kitware.com
>>>>>> [mailto:rtk-users at public.kitware.com],
>>>>>>>>>> saimahesh.m at panaceamedical.com
>>>>>>>>>>> Subject: Re: [Rtk-users] Reg: CBCT 1k reconstruction error
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>>> Dear Zahid,
>>>>>>>>>>>> There is no screenshot so I'm guessing that the problem is a CUDA
>>>>>>>> memory
>>>>>>>>>>>> error. If you use the command line tool rtkfdk, you should first
>>>> make
>>>>>>>>>> sure
>>>>>>>>>>>> that you use the --lowmem option to stream the projection images.
>>>> If
>>>>>> it
>>>>>>>>>> is
>>>>>>>>>>>> no sufficient, you can split your volume using the --divisions.
>> The
>>>>>>>>>>>> corresponding C++ code can be found in rtkfdk.cxx.
>>>>>>>>>>>> Is Panacea Medical using RTK? We are happy to advertise the use
>> of
>>>>>> RTK
>>>>>>>> on
>>>>>>>>>>>> our case studies webpage
>>>>>>>>>> <http://openrtk.org/RTK/project/casestudies.html>.
>>>>>>>>>>>> Simon
>>>>>>>>>>>>
>>>>>>>>>>>> On Sat, Oct 15, 2016 at 11:12 AM, Zahid Hasan Ansari <
>>>>>>>>>>>> zahidhasan.a at panaceamedical.com> wrote:
>>>>>>>>>>>>
>>>>>>>>>>>>>> Dear Sir\Madam,
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> I am unable to do 1024x1024x1024
>>>> reconstruction
>>>>>>>> using
>>>>>>>>>> RTK
>>>>>>>>>>>>>> CudaFDKConeBeamReconstructionFilter. Please check the screen
>>>> shoot
>>>>>> of
>>>>>>>>>>>>> the
>>>>>>>>>>>>>> same.
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> I am using the following items given below.
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> 1. RTK - RTK version 1.3
>>>>>>>>>>>>>> 2. ITK - ITK version 4.7
>>>>>>>>>>>>>> 3. Visual Studio 2013 win 64-bit console
>>>>>> application
>>>>>>>>>>>>>> 4. GeForce GTX TITAN X GPU.
>>>>>>>>>>>>>> 5. CUDA 7.
>>>>>>>>>>>>>> 6. CMake version 3.4.3.
>>>>>>>>>>>>>> 7. Windows 7 64-bit OS.
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> Please provide the solution of this and let
>> me
>>>>>> know
>>>>>>>> if
>>>>>>>>>>>> any
>>>>>>>>>>>>>> other clarification is required.
>>>>>>>>>>>>>>
>>>>>>>>>>>>>>
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> Thanks & Regards
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> Zahid Hasan Ansari
>>>>>>>>>>>>>> Senior Design Engineer
>>>>>>>>>>>>>> Mobile No. +91-9738379729
>>>>>>>>>>>>>>
>>>>>>>>>>>>>> Panacea Medical Technologies Pvt. Ltd.
>>>>>>>>>>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP
>> Area
>>>>>>>> Phase
>>>>>>>>>>>>> 1,
>>>>>>>>>>>>>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>>>>>>>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB
>> Industrial
>>>>>>>> Area,
>>>>>>>>>>>>>> Malur - 563130. Kolar District. INDIA.
>>>>>>>>>>>>>> Tel : +91 80 4242 8700 / 2845 4785 | Fax : + 91 42428710 | Url
>> :
>>>>>>>>>>>>>> http://www.panaceamedical.com
>>>>>>>>>>>>>> Bangalore - India.
>>>>>>>>>>>>> ________________________________________
>>>>>>>>>>>>> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
>>>>>>>>>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP
>> Area
>>>>>>>> Phase
>>>>>>>>>>>>> 1, Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>>>>>>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB
>> Industrial
>>>>>>>> Area,
>>>>>>>>>>>>> Malur - 563130. Kolar District. INDIA.
>>>>>>>>>>>>>
>>>>>>>>>>>>> Tel : +91 80 4242 8700 / 2845 4785
>>>>>>>>>>>>> Fax : + 91 80 42428710
>>>>>>>>>>>>> Url : http://www.panaceamedical.com
>>>>>>>>>>>>> ____________________________________________________________
>>>>>>>>>>>>> ________________
>>>>>>>>>>>>> PMT EMAIL DISCLAIMER:
>>>>>>>>>>>>>
>>>>>>>>>>>>> This email and any files transmitted with it are confidential
>> and
>>>>>>>>>> intended
>>>>>>>>>>>>> solely for the use of the individual or entity to whom they are
>>>>>>>>>> addressed.
>>>>>>>>>>>>> If you have received this email in error please notify the
>> system
>>>>>>>>>> manager.
>>>>>>>>>>>>> Please note that any views or opinions presented in this email
>> are
>>>>>>>>>> solely
>>>>>>>>>>>>> those of the author and do not necessarily represent those of
>> the
>>>>>>>>>> company.
>>>>>>>>>>>>> Finally, the recipient should check this email and any
>> attachments
>>>>>> for
>>>>>>>>>> the
>>>>>>>>>>>>> presence of viruses. The company accepts no liability for any
>>>> damage
>>>>>>>>>>>>> caused by any virus transmitted by this email.
>>>>>>>>>>>>>
>>>>>>>>>>>>>
>>>>>>>>>>>>> _______________________________________________
>>>>>>>>>>>>> Rtk-users mailing list
>>>>>>>>>>>>> Rtk-users at public.kitware.com
>>>>>>>>>>>>> http://public.kitware.com/mailman/listinfo/rtk-users
>>>>>>>>>>>>>
>>>>>>>>>>> ________________________________________
>>>>>>>>>>> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
>>>>>>>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area
>>>>>> Phase
>>>>>>>> 1,
>>>>>>>>>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>>>>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial
>>>>>> Area,
>>>>>>>>>> Malur - 563130. Kolar District. INDIA.
>>>>>>>>>>> Tel : +91 80 4242 8700 / 2845 4785
>>>>>>>>>>> Fax : + 91 80 42428710
>>>>>>>>>>> Url : http://www.panaceamedical.com
>>>>>>>>>>>
>> ____________________________________________________________________________
>>>>>>>>>>> PMT EMAIL DISCLAIMER:
>>>>>>>>>>>
>>>>>>>>>>> This email and any files transmitted with it are confidential and
>>>>>>>> intended
>>>>>>>>>>> solely for the use of the individual or entity to whom they are
>>>>>>>> addressed.
>>>>>>>>>>> If you have received this email in error please notify the system
>>>>>>>> manager.
>>>>>>>>>>> Please note that any views or opinions presented in this email are
>>>>>>>> solely
>>>>>>>>>>> those of the author and do not necessarily represent those of the
>>>>>>>> company.
>>>>>>>>>>> Finally, the recipient should check this email and any attachments
>>>> for
>>>>>>>> the
>>>>>>>>>>> presence of viruses. The company accepts no liability for any
>> damage
>>>>>>>>>>> caused by any virus transmitted by this email.
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>>
>>>>>>>>>>> _______________________________________________
>>>>>>>>>>> Rtk-users mailing list
>>>>>>>>>>> Rtk-users at public.kitware.com
>>>>>>>>>>> http://public.kitware.com/mailman/listinfo/rtk-users
>>>>>>>>> ________________________________________
>>>>>>>>> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
>>>>>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area
>>>> Phase
>>>>>> 1,
>>>>>>>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial
>>>> Area,
>>>>>>>> Malur - 563130. Kolar District. INDIA.
>>>>>>>>> Tel : +91 80 4242 8700 / 2845 4785
>>>>>>>>> Fax : + 91 80 42428710
>>>>>>>>> Url : http://www.panaceamedical.com
>>>>>>>>>
>> ____________________________________________________________________________
>>>>>>>>> PMT EMAIL DISCLAIMER:
>>>>>>>>>
>>>>>>>>> This email and any files transmitted with it are confidential and
>>>>>> intended
>>>>>>>>> solely for the use of the individual or entity to whom they are
>>>>>> addressed.
>>>>>>>>> If you have received this email in error please notify the system
>>>>>> manager.
>>>>>>>>> Please note that any views or opinions presented in this email are
>>>>>> solely
>>>>>>>>> those of the author and do not necessarily represent those of the
>>>>>> company.
>>>>>>>>> Finally, the recipient should check this email and any attachments
>> for
>>>>>> the
>>>>>>>>> presence of viruses. The company accepts no liability for any damage
>>>>>>>>> caused by any virus transmitted by this email.
>>>>>>>>>
>>>>>>> ________________________________________
>>>>>>> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
>>>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area
>> Phase
>>>> 1,
>>>>>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial
>> Area,
>>>>>> Malur - 563130. Kolar District. INDIA.
>>>>>>> Tel : +91 80 4242 8700 / 2845 4785
>>>>>>> Fax : + 91 80 42428710
>>>>>>> Url : http://www.panaceamedical.com
>>>>>>>
>> ____________________________________________________________________________
>>>>>>> PMT EMAIL DISCLAIMER:
>>>>>>>
>>>>>>> This email and any files transmitted with it are confidential and
>>>> intended
>>>>>>> solely for the use of the individual or entity to whom they are
>>>> addressed.
>>>>>>> If you have received this email in error please notify the system
>>>> manager.
>>>>>>> Please note that any views or opinions presented in this email are
>>>> solely
>>>>>>> those of the author and do not necessarily represent those of the
>>>> company.
>>>>>>> Finally, the recipient should check this email and any attachments for
>>>> the
>>>>>>> presence of viruses. The company accepts no liability for any damage
>>>>>>> caused by any virus transmitted by this email.
>>>>>>>
>>>>>>>
>>>>>>>
>>>>> ________________________________________
>>>>> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
>>>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area Phase
>> 1,
>>>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial Area,
>>>> Malur - 563130. Kolar District. INDIA.
>>>>> Tel : +91 80 4242 8700 / 2845 4785
>>>>> Fax : + 91 80 42428710
>>>>> Url : http://www.panaceamedical.com
>>>>>
>> ____________________________________________________________________________
>>>>> PMT EMAIL DISCLAIMER:
>>>>>
>>>>> This email and any files transmitted with it are confidential and
>> intended
>>>>> solely for the use of the individual or entity to whom they are
>> addressed.
>>>>> If you have received this email in error please notify the system
>> manager.
>>>>> Please note that any views or opinions presented in this email are
>> solely
>>>>> those of the author and do not necessarily represent those of the
>> company.
>>>>> Finally, the recipient should check this email and any attachments for
>> the
>>>>> presence of viruses. The company accepts no liability for any damage
>>>>> caused by any virus transmitted by this email.
>>>>>
>>>>
>>> ________________________________________
>>> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
>>> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area Phase 1,
>> Whitefield, Bangalore - 560 066, Karnataka, INDIA
>>> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial Area,
>> Malur - 563130. Kolar District. INDIA.
>>> Tel : +91 80 4242 8700 / 2845 4785
>>> Fax : + 91 80 42428710
>>> Url : http://www.panaceamedical.com
>>>
>> ____________________________________________________________________________
>>> PMT EMAIL DISCLAIMER:
>>>
>>> This email and any files transmitted with it are confidential and intended
>>> solely for the use of the individual or entity to whom they are addressed.
>>> If you have received this email in error please notify the system manager.
>>> Please note that any views or opinions presented in this email are solely
>>> those of the author and do not necessarily represent those of the company.
>>> Finally, the recipient should check this email and any attachments for the
>>> presence of viruses. The company accepts no liability for any damage
>>> caused by any virus transmitted by this email.
>>>
>>>
>>>
>>
>>
> ________________________________________
> PANACEA MEDICAL TECHNOLOGIES PVT. LTD.
> Head Office: Plot #119, GF, Envision Technology Center, EPIP Area Phase 1, Whitefield, Bangalore - 560 066, Karnataka, INDIA
> Manufacturing Unit: Plot #87 A, Phase -II, Malur KIADB Industrial Area, Malur - 563130. Kolar District. INDIA.
>
> Tel : +91 80 4242 8700 / 2845 4785
> Fax : + 91 80 42428710
> Url : http://www.panaceamedical.com
> ____________________________________________________________________________
> PMT EMAIL DISCLAIMER:
>
> This email and any files transmitted with it are confidential and intended
> solely for the use of the individual or entity to whom they are addressed.
> If you have received this email in error please notify the system manager.
> Please note that any views or opinions presented in this email are solely
> those of the author and do not necessarily represent those of the company.
> Finally, the recipient should check this email and any attachments for the
> presence of viruses. The company accepts no liability for any damage
> caused by any virus transmitted by this email.
>
>
>
>
>
More information about the Rtk-users
mailing list