Hi Luis,
thanks for your response, but as I said, I could fix the problem.
The problem is, that the Serieswriter ignores the "Modality rescale
Intercept Element"
and "Modality Rescale Slope Element" of the dicom header, but the reader
seems not
to ignore it when you load the serie again.
So, I just change the dicom - tag for the slope ("0028|1052") before I
write the serie,
and it works fine then.
Regards,
Markus
Luis Ibanez schrieb:
>
> Hi Markus,
>
> Please post to the mailing list the code that you
> used for writing the DicomSlices.
>
>
> Thanks
>
>
> Luis
>
>
>
> ======================
> Markus Weigert wrote:
>
>> Hi,
>>
>> today I encountered a strange problem with dicom ImageIO:
>>
>> After reading a dicom series of CT - files and displaying the volume
>> with
>> GLTwoImageSliceView (from insight applications), everything is fine.
>> The range of the values is -1024 up to 3071, just as it should be.
>>
>> But after writing the serie again to dicom slices and reading these
>> slices,
>> the range is now -2048 to 2047, so there seems to be a shift.
>> I don't understand this, because I didn't change anything in the
>> MetaDataDictionaryArray.
>>
>> Has anybody an idea what could cause this shifting of image values?
>>
>> Regards,
>> Markus
>>