Hi Luis,

Thank you for your reply.

I think my previous mail was not as clear as it should.

In fact, the following code is in the igstkTracker.cxx file, at line 583 :
>
>>   TransformType identityTransform;
>>   identityTransform.SetToIdentity( 
>> igstk::TimeStamp::GetLongestPossibleTime() );
>>  
>>   m_TrackerToolToBeAttached->RequestSetTransformAndParent(
>>                               identityTransform, this );
It is called when a TrackerTool is attached to the Tracker.
I proposed to set the validity time to 0 instead of the 
LongestPossibleTime, in order to not display the tracker tool if its 
transformation is never updated.

Vincent.


Luis Ibanez a écrit :
> Hi Vincent,
>
> The recomended way of setting the validity time of your tracker is
> as a multiple of the sampling period of the hardware tracking device.
>
> For example,
>
>    If you hardware Tracker is capable of providing positions
>    at 20Hz, its sampling period will be 1/20 seconds (0.05s).
>
>    You want to setup the rate of the software igstkTracker
>    class to 20Hz or less, and internally, in the software
>    class, you set the validity time of a Transform to 1/f
>    where F is the frequency of the Tracker.
>
>
> That is, if you are sampling Transforms at 20Hz (e.g. 20 per second)
> it is reasonable to assume that the validity time of each Transform
> is only 0.05 seconds, because by then, a new Transform should be
> made available.
>
> If you want to be in the safe side, you may want to increase that
> validity time to 2/f or 3/f, ( 100 ms, or 150ms) to avoid accidental
> flickering of the objects in the scene due to small timing delays
> on the delivery of the Transform.
>
> To summarize:
>
>    Your Tracker *should not* use the LongestPossibleTime as the
>    validity time.
>
>    The recommended validity time is 2/Frequency or 3/Frequency.
>
>
>
>   Regards,
>
>
>      Luis
>
>
> ------------------------
> Vincent Gratsac wrote:
>> Hi all,
>>
>> I developed a tracker class which inherits igstk::Tracker and 
>> communicates with the tracker that we are using in operating room.
>> I noted that if the user adds a tool to this tracker, and if this to 
>> is never visible, then its transformation will never been updated. 
>> But the tracker tool representation will be always visible !
>> This comes from the initialization of the tool transformation when it 
>> is attached to the tracker (l. 583):
>>
>>>   TransformType identityTransform;
>>>   identityTransform.SetToIdentity( 
>>> igstk::TimeStamp::GetLongestPossibleTime() );
>>>  
>>>   m_TrackerToolToBeAttached->RequestSetTransformAndParent(
>>>                               identityTransform, this );
>>
>>
>> So if the TrackerTool's transformation is never modified (because 
>> tool is never visible), it will always be an identity transformation 
>> with an infinite validity.
>> I tried to replace
>>
>>> identityTransform.SetToIdentity( 
>>> igstk::TimeStamp::GetLongestPossibleTime() );
>>
>>
>> by
>>
>>> identityTransform.SetToIdentity( 0 );
>>
>>
>> and it works. Now, an attached TrackerTool will be displayed only 
>> when a valid transformation is set by the 
>> UpdateStatusSuccessProcessing() method.
>>
>> Do you think this is a correct way to do ?
>>
>> Vincent
>>
>> P.S : Modified igstkTracker.cxx file is attached.
>>
>>
>>
>> ------------------------------------------------------------------------
>>
>> /*========================================================================= 
>>
>>
>>   Program:   Image Guided Surgery Software Toolkit
>>   Module:    $RCSfile: igstkTracker.cxx,v $
>>   Language:  C++
>>   Date:      $Date: 2008/02/11 22:12:59 $
>>   Version:   $Revision: 1.58 $
>>
>>   Copyright (c) ISC  Insight Software Consortium.  All rights reserved.
>>   See IGSTKCopyright.txt or http://www.igstk.org/copyright.htm for 
>> details.
>>
>>      This software is distributed WITHOUT ANY WARRANTY; without even
>>      the implied warranty of MERCHANTABILITY or FITNESS FOR A PARTICULAR
>>      PURPOSE.  See the above copyright notices for more information.
>>
>> =========================================================================*/ 
>>
>>
>> #if defined(_MSC_VER)
>> //Warning about: identifier was truncated to '255' characters in the 
>> debug // information (MVC6.0 Debug)
>> #pragma warning( disable : 4786 )
>> #endif
>>
>> #include "igstkTracker.h"
>>
>> namespace igstk
>> {
>>
>>
>> /** Constructor */
>> Tracker::Tracker(void) :  m_StateMachine( this ) {
>>   /** Coordinate system interface */
>>   igstkCoordinateSystemClassInterfaceConstructorMacro();
>>
>>   // Set the state descriptors
>>   igstkAddStateMacro( Idle );   igstkAddStateMacro( 
>> AttemptingToEstablishCommunication );   igstkAddStateMacro( 
>> AttemptingToCloseCommunication);   igstkAddStateMacro( 
>> AttemptingToAttachTrackerTool );
>>   igstkAddStateMacro( TrackerToolAttached );
>>   igstkAddStateMacro( CommunicationEstablished );
>>   igstkAddStateMacro( AttemptingToTrack );   igstkAddStateMacro( 
>> AttemptingToStopTracking);   igstkAddStateMacro( Tracking );   
>> igstkAddStateMacro( AttemptingToUpdate );     // Set the input 
>> descriptors
>>   igstkAddInputMacro( EstablishCommunication);
>>   igstkAddInputMacro( AttachTrackerTool);
>>   igstkAddInputMacro( StartTracking);   igstkAddInputMacro( 
>> UpdateStatus);   igstkAddInputMacro( StopTracking);   
>> igstkAddInputMacro( Reset);   igstkAddInputMacro( 
>> CloseCommunication);   igstkAddInputMacro( Success);   
>> igstkAddInputMacro( Failure);   igstkAddInputMacro( ValidFrequency);
>>   // Programming the state machine transitions:
>>
>>   // Transitions from the Idle
>>   igstkAddTransitionMacro( Idle,
>>                            EstablishCommunication,
>>                            AttemptingToEstablishCommunication,
>>                            AttemptToOpen );
>>   igstkAddTransitionMacro( Idle,
>>                            StartTracking,
>>                            Idle,
>>                            ReportInvalidRequest );
>>   igstkAddTransitionMacro( Idle,
>>                            StopTracking,
>>                            Idle,
>>                            ReportInvalidRequest );
>>   igstkAddTransitionMacro( Idle,
>>                            AttachTrackerTool,
>>                            Idle,
>>                            ReportInvalidRequest );
>>   igstkAddTransitionMacro( Idle,
>>                            UpdateStatus,
>>                            Idle,
>>                            ReportInvalidRequest );
>>   igstkAddTransitionMacro( Idle,
>>                            Reset,
>>                            Idle,
>>                            ReportInvalidRequest );
>>   igstkAddTransitionMacro( Idle,
>>                            CloseCommunication,
>>                            Idle,
>>                            ReportInvalidRequest );
>>   igstkAddTransitionMacro( Idle,
>>                            ValidFrequency,
>>                            Idle,
>>                            SetFrequency );
>>
>>   // Transitions from the AttemptingToEstablishCommunication
>>   igstkAddTransitionMacro( AttemptingToEstablishCommunication,
>>                            Success,
>>                            CommunicationEstablished,
>>                            CommunicationEstablishmentSuccess );
>>
>>   igstkAddTransitionMacro( AttemptingToEstablishCommunication,
>>                            Failure,
>>                            Idle,
>>                            CommunicationEstablishmentFailure );
>>
>>   igstkAddTransitionMacro( AttemptingToEstablishCommunication,
>>                            ValidFrequency,
>>                            AttemptingToEstablishCommunication,
>>                            ReportInvalidRequest );
>>
>>   // Transitions from CommunicationEstablished
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            AttachTrackerTool,
>>                            AttemptingToAttachTrackerTool,
>>                            AttemptToAttachTrackerTool );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            StartTracking,
>>                            AttemptingToTrack,
>>                            AttemptToStartTracking );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            CloseCommunication,
>>                            AttemptingToCloseCommunication,
>>                            CloseFromCommunicatingState );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            Reset,
>>                            CommunicationEstablished,
>>                            ResetFromCommunicatingState );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            StopTracking,
>>                            CommunicationEstablished,
>>                            ReportInvalidRequest );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            EstablishCommunication,
>>                            CommunicationEstablished,
>>                            ReportInvalidRequest );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            UpdateStatus,
>>                            CommunicationEstablished,
>>                            ReportInvalidRequest );
>>
>>   igstkAddTransitionMacro( CommunicationEstablished,
>>                            ValidFrequency,
>>                            CommunicationEstablished,
>>                            SetFrequency );
>>
>>   // Transitions from AttemptingToAttachTrackerTool
>>   igstkAddTransitionMacro( AttemptingToAttachTrackerTool,
>>                            Success,
>>                            TrackerToolAttached,
>>                            AttachingTrackerToolSuccess );
>>
>>   igstkAddTransitionMacro( AttemptingToAttachTrackerTool,
>>                            Failure,
>>                            CommunicationEstablished,
>>                            AttachingTrackerToolFailure );
>>
>>   igstkAddTransitionMacro( AttemptingToAttachTrackerTool,
>>                            ValidFrequency,
>>                            AttemptingToAttachTrackerTool,
>>                            ReportInvalidRequest );
>>
>>   // Transitions from TrackerToolAttached
>>   igstkAddTransitionMacro( TrackerToolAttached,
>>                            StartTracking,
>>                            AttemptingToTrack,
>>                            AttemptToStartTracking );
>>
>>   igstkAddTransitionMacro( TrackerToolAttached,
>>                            AttachTrackerTool,
>>                            AttemptingToAttachTrackerTool,
>>                            AttemptToAttachTrackerTool );
>>
>>   igstkAddTransitionMacro( TrackerToolAttached,
>>                            ValidFrequency,
>>                            TrackerToolAttached,
>>                            SetFrequency );
>>
>>   igstkAddTransitionMacro( TrackerToolAttached,
>>                            StopTracking,
>>                            TrackerToolAttached,
>>                            ReportInvalidRequest );
>>
>>   igstkAddTransitionMacro( TrackerToolAttached,
>>                            CloseCommunication,
>>                            AttemptingToCloseCommunication,
>>                            CloseFromCommunicatingState );
>>
>>   // Transitions from AttemptingToTrack
>>   igstkAddTransitionMacro( AttemptingToTrack,
>>                            Success,
>>                            Tracking,
>>                            StartTrackingSuccess );
>>
>>   igstkAddTransitionMacro( AttemptingToTrack,
>>                            Failure,
>>                            CommunicationEstablished,
>>                            StartTrackingFailure );
>>
>>   igstkAddTransitionMacro( AttemptingToTrack,
>>                            ValidFrequency,
>>                            AttemptingToTrack,
>>                            ReportInvalidRequest );
>>
>>   // Transitions from Tracking
>>   igstkAddTransitionMacro( Tracking,
>>                            UpdateStatus,
>>                            AttemptingToUpdate,
>>                            AttemptToUpdateStatus );
>>
>>   igstkAddTransitionMacro( Tracking,
>>                            StopTracking,
>>                            AttemptingToStopTracking,
>>                            AttemptToStopTracking );
>>
>>   igstkAddTransitionMacro( Tracking,
>>                            Reset,
>>                            CommunicationEstablished,
>>                            ResetFromTrackingState );
>>
>>   igstkAddTransitionMacro( Tracking,
>>                            CloseCommunication,
>>                            AttemptingToCloseCommunication,
>>                            CloseFromTrackingState );
>>
>>   igstkAddTransitionMacro( Tracking,
>>                            ValidFrequency,
>>                            Tracking,
>>                            ReportInvalidRequest );
>>
>>   // Transitions from AttemptingToUpdate
>>   igstkAddTransitionMacro( AttemptingToUpdate,
>>                            Success,
>>                            Tracking,
>>                            UpdateStatusSuccess );
>>
>>   igstkAddTransitionMacro( AttemptingToUpdate,
>>                            Failure,
>>                            Tracking,
>>                            UpdateStatusFailure );
>>
>>   igstkAddTransitionMacro( AttemptingToUpdate,
>>                            ValidFrequency,
>>                            AttemptingToUpdate,
>>                            ReportInvalidRequest );
>>
>>   // Transitions from AttemptingToStopTracking
>>   igstkAddTransitionMacro( AttemptingToStopTracking,
>>                            Success,
>>                            CommunicationEstablished,
>>                            StopTrackingSuccess );
>>
>>   igstkAddTransitionMacro( AttemptingToStopTracking,
>>                            Failure,
>>                            Tracking,
>>                            StopTrackingFailure );
>>
>>   igstkAddTransitionMacro( AttemptingToStopTracking,
>>                            ValidFrequency,
>>                            AttemptingToStopTracking,
>>                            ReportInvalidRequest );
>>
>>   igstkAddTransitionMacro( AttemptingToCloseCommunication,
>>                            Success,
>>                            Idle,
>>                            CloseCommunicationSuccess );
>>
>>   igstkAddTransitionMacro( AttemptingToCloseCommunication,
>>                            Failure,
>>                            CommunicationEstablished,
>>                            CloseCommunicationFailure );
>>
>>   igstkAddTransitionMacro( AttemptingToCloseCommunication,
>>                            ValidFrequency,
>>                            AttemptingToCloseCommunication,
>>                            ReportInvalidRequest );
>>
>>   // Select the initial state of the state machine
>>   igstkSetInitialStateMacro( Idle );
>>
>>   // Finish the programming and get ready to run
>>   m_StateMachine.SetReadyToRun();
>>
>>   // Create a PulseGenerator object.    m_PulseGenerator = 
>> PulseGenerator::New();
>>
>>   m_PulseObserver = ObserverType::New();
>>   m_PulseObserver->SetCallbackFunction( this, & 
>> Tracker::RequestUpdateStatus );
>>   m_PulseGenerator->AddObserver( PulseEvent(), m_PulseObserver );
>>
>>   // This is update rate for sending tracking information to the
>>   // spatial objects, it should be set to at least 30 Hz
>>   const double DEFAULT_REFRESH_RATE = 30.0;
>>   m_PulseGenerator->RequestSetFrequency( DEFAULT_REFRESH_RATE );
>>
>>   // This is the time period for which transformation should be
>>   // considered valid.  After this time, they expire.  This time
>>   // is in milliseconds.
>>   const TimePeriodType DEFAULT_VALIDITY_TIME = 400;
>>   m_ValidityTime = DEFAULT_VALIDITY_TIME;
>>
>>   // By default, the reference is not used
>>   m_ApplyingReferenceTool = false;
>>
>>   m_ConditionNextTransformReceived = itk::ConditionVariable::New();
>>   m_Threader = itk::MultiThreader::New();
>>   m_ThreadingEnabled = false;
>>   m_TrackingThreadStarted = false;
>> }
>>
>> /** Destructor */
>> Tracker::~Tracker(void)
>> {
>> }
>>
>> /** This method sets the reference tool. */
>> void Tracker::RequestSetReferenceTool( TrackerToolType * trackerTool )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestSetReferenceTool 
>> called ...\n");
>>   // connect the reference tracker tool the tracker   TransformType 
>> identityTransform;
>>   identityTransform.SetToIdentity(                     
>> igstk::TimeStamp::GetLongestPossibleTime() );
>>     if( trackerTool != NULL )
>>     {
>>     // check if it is already attached to the tracker
>>     typedef TrackerToolsContainerType::iterator InputIterator;
>>     InputIterator toolItr =                 m_TrackerTools.find( 
>> trackerTool->GetTrackerToolIdentifier() );
>>
>>     if( toolItr != m_TrackerTools.end() )
>>       {
>>       m_ApplyingReferenceTool = true;
>>       m_ReferenceTool = trackerTool;
>>       m_ReferenceTool->RequestDetachFromParent();
>>
>>       //VERY IMPORTANT: Make the reference tracker tool the parent of 
>> the tracker
>>       this->RequestSetTransformAndParent( identityTransform, 
>> trackerTool );
>>       }
>>     else
>>       {
>>       igstkLogMacro( CRITICAL, "Request to use a tracker tool as a 
>> reference"
>>       << "has failed. The tracker tool is not attached to the tracker 
>> ");
>>       }
>>    }
>> }
>>
>> /** The "RequestOpen" method attempts to open communication with the
>>  *  tracking device. */
>> void Tracker::RequestOpen( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestOpen called...\n");
>>   igstkPushInputMacro( EstablishCommunication );
>>   this->m_StateMachine.ProcessInputs();
>> }
>>
>>
>> /** The "RequestClose" method closes communication with the device. */
>> void Tracker::RequestClose( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestClose called ...\n");
>>   igstkPushInputMacro( CloseCommunication );
>>   m_StateMachine.ProcessInputs();
>> }
>>
>> /** The "RequestReset" tracker method should be used to the tracker
>>  * to some defined default state. */
>> void Tracker::RequestReset( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestReset called ...\n");
>>   igstkPushInputMacro( Reset );
>>   m_StateMachine.ProcessInputs();
>> }
>>
>>
>> /** The "RequestStartTracking" method readies the tracker for 
>> tracking the
>>  *  tools connected to the tracker. */
>> void Tracker::RequestStartTracking( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestStartTracking called 
>> ...\n");
>>   igstkPushInputMacro( StartTracking );
>>   m_StateMachine.ProcessInputs();
>> }
>>
>>
>> /** The "RequestStopTracking" stops tracker from tracking the tools. */
>> void Tracker::RequestStopTracking( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestStopTracking called 
>> ...\n");
>>   igstkPushInputMacro( StopTracking );
>>   m_StateMachine.ProcessInputs();
>> }
>>
>>
>> /** The "RequestUpdateStatus" method is used for updating the status 
>> of  *  ports and tools when the tracker is in tracking state. */
>> void Tracker::RequestUpdateStatus( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestUpdateStatus called 
>> ...\n");
>>   igstkPushInputMacro( UpdateStatus );
>>   m_StateMachine.ProcessInputs();
>> }
>>
>>
>> /** The "RequestSetFrequency" method is used for defining the rate at 
>> which  * Transforms are queried from the Tracker device */
>> void Tracker::RequestSetFrequency( double frequencyInHz )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::RequestSetFrequency called 
>> ...\n");
>>   if( frequencyInHz > 0.0 )
>>     {
>>     this->m_FrequencyToBeSet = frequencyInHz;
>>     igstkPushInputMacro( ValidFrequency );
>>     m_StateMachine.ProcessInputs();
>>     }
>> }
>>
>> /** The "AttemptToOpen" method attempts to open communication with a
>>  *  tracking device. */
>> void Tracker::AttemptToOpenProcessing( void )
>> {
>>   igstkLogMacro( DEBUG,                  
>> "igstk::Tracker::AttemptToOpenProcessing called ...\n");
>>
>>   ResultType result = this->InternalOpen();
>>     m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>>
>> /** Post-processing after communication setup has been successful. */ 
>> void Tracker::CommunicationEstablishmentSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG,     
>> "igstk::Tracker::CommunicationEstablishmentSuccessProcessing called 
>> ...\n");
>>
>>   this->InvokeEvent( TrackerOpenEvent() );
>> }
>>
>>
>> /** Post-processing after communication setup has failed. */ void 
>> Tracker::CommunicationEstablishmentFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::"
>>                  "CommunicationEstablishmentFailureProcessing called 
>> ...\n");
>>
>>   this->InvokeEvent( TrackerOpenErrorEvent() );
>> }
>>
>> /** The Reset methods force the tracker to the
>>  *  CommunicationEstablished state */
>> void Tracker::ResetFromTrackingStateProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::"
>>                  "ResetFromTrackingStateProcessing() called ...\n");
>>   // leaving TrackingState, going to CommunicationEstablishedState
>>   this->ExitTrackingStateProcessing();
>>   this->ResetFromToolsActiveStateProcessing();
>> }
>>
>> /** The Reset methods force the tracker to the
>>  *  CommunicationEstablished  state */
>> void Tracker::ResetFromToolsActiveStateProcessing( void )
>> {
>>   igstkLogMacro( DEBUG,          
>> "igstk::Tracker::ResetFromToolsActiveStateProcessing() called ...\n");
>>   this->ResetFromCommunicatingStateProcessing();
>> }
>>
>> /** The Reset methods force the tracker to the
>>  *  CommunicationEstablished state */
>> void Tracker::ResetFromCommunicatingStateProcessing( void )
>> {
>>   ResultType result = this->InternalReset();
>>
>>   if( result == SUCCESS )
>>     {
>>     igstkLogMacro( DEBUG, "igstk::Tracker::InternalReset succeeded 
>> ...\n");
>>     }
>>   else if( result == FAILURE )
>>     {
>>     igstkLogMacro( DEBUG, "igstk::Tracker::InternalReset failed ...\n");
>>     }
>> }
>>
>> /** Post-processing after ports and tools setup has been successful. 
>> */ void Tracker::ToolsActivationSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::ToolsActivationSuccessProcessing "
>>                  "called ...\n");
>>
>>   this->InvokeEvent( TrackerInitializeEvent() );
>> }
>>
>> /** Post-processing after ports and tools setup has failed. */ void 
>> Tracker::ToolsActivationFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::ToolsActivationFailureProcessing "
>>                  "called ...\n");
>>
>>   this->InvokeEvent( TrackerInitializeErrorEvent() );
>> }
>>
>> /** The "AttemptToStartTracking" method attempts to start tracking. */
>> void Tracker::AttemptToStartTrackingProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::AttemptToStartTrackingProcessing  "
>>                  "called ...\n");
>>
>>   ResultType result = this->InternalStartTracking();
>>     m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>> /** Post-processing after start tracking has been successful. */ void 
>> Tracker::StartTrackingSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::StartTrackingSuccessProcessing "
>>                  "called ...\n");
>>
>>   // Report to all the tracker tools that tracking has been started
>>   typedef TrackerToolsContainerType::iterator  InputConstIterator;
>>
>>   InputConstIterator inputItr = m_TrackerTools.begin();
>>   InputConstIterator inputEnd = m_TrackerTools.end();
>>
>>   while( inputItr != inputEnd )
>>     {
>>     (inputItr->second)->RequestReportTrackingStarted();
>>     ++inputItr;
>>     }
>>
>>
>>   // going from AttemptingToTrackState to TrackingState
>>   this->EnterTrackingStateProcessing();
>>
>>   this->InvokeEvent( TrackerStartTrackingEvent() );
>> }
>>
>> /** Post-processing after start tracking has failed. */ void 
>> Tracker::StartTrackingFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::StartTrackingFailureProcessing "
>>                  "called ...\n");
>>
>>   this->InvokeEvent( TrackerStartTrackingErrorEvent() );
>> }
>>
>> /** Post-processing after attaching a tracker tool to the tracker
>>  *  has been successful. */ void 
>> Tracker::AttachingTrackerToolSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::AttachingTrackerToolSuccessProcessing "
>>                  "called ...\n");
>>
>>   m_TrackerTools[ 
>> m_TrackerToolToBeAttached->GetTrackerToolIdentifier() ] 
>>                                    = m_TrackerToolToBeAttached;
>>   // report to the tracker tool that the attachment has been   // 
>> successful
>>   
>> m_TrackerToolToBeAttached->RequestReportSuccessfulTrackerToolAttachment(); 
>>
>>
>>   // Add the tracker tool to the internal data containers
>>   this->AddTrackerToolToInternalDataContainers( 
>> m_TrackerToolToBeAttached );
>>
>>   //connect the tracker tool coordinate system to the tracker
>>   //system. By default, make the tracker coordinate system to   //be 
>> a parent of the tracker tool coordinate system
>>   //If a reference tracker tool is specified, the reference
>>   //tracker tool will become the parent of all the tracker tools.
>>   TransformType identityTransform;
>>   identityTransform.SetToIdentity( 0 );
>>   m_TrackerToolToBeAttached->RequestSetTransformAndParent( 
>>                               identityTransform, this );
>>     this->InvokeEvent( AttachingTrackerToolToTrackerEvent() );
>> }
>>
>> /** Post-processing after attaching a tracker tool to the tracker
>>  *  has failed. */ void 
>> Tracker::AttachingTrackerToolFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::AttachingTrackerToolFailureProcessing "
>>                  "called ...\n");
>>
>>   // report to the tracker tool that the attachment has failed
>>   m_TrackerToolToBeAttached->RequestReportFailedTrackerToolAttachment();
>>
>>   this->InvokeEvent( AttachingTrackerToolToTrackerErrorEvent() );
>> }
>>  
>> /** The "AttemptToStopTracking" method attempts to stop tracking. */
>> void Tracker::AttemptToStopTrackingProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::AttemptToStopTrackingProcessing "
>>                         "called ...\n");
>>   // leaving TrackingState, going to AttemptingToStopTrackingState
>>   this->ExitTrackingStateProcessing();
>>
>>   ResultType result = this->InternalStopTracking();
>>     m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>>
>> /** Post-processing after stop tracking has been successful. */ void 
>> Tracker::StopTrackingSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::StopTrackingSuccessProcessing "
>>                  "called ...\n");
>>
>>   // Report to all the tracker tools that tracking has been stopped
>>   typedef TrackerToolsContainerType::iterator  InputConstIterator;
>>
>>   InputConstIterator inputItr = m_TrackerTools.begin();
>>   InputConstIterator inputEnd = m_TrackerTools.end();
>>
>>   while( inputItr != inputEnd )
>>     {
>>     (inputItr->second)->RequestReportTrackingStopped();
>>     ++inputItr;
>>     }
>>
>>   this->InvokeEvent( TrackerStopTrackingEvent() );  }
>>
>> /** Post-processing after start tracking has failed. */ void 
>> Tracker::StopTrackingFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::StopTrackingFailureProcessing "
>>                  "called ...\n");
>>   // going from AttemptingToStopTrackingState to TrackingState
>>   this->EnterTrackingStateProcessing();
>>
>>   this->InvokeEvent( TrackerStopTrackingErrorEvent() );  }
>>
>> /** Needs to be called every time when entering tracking state. */ 
>> void Tracker::EnterTrackingStateProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::EnterTrackingStateProcessing "
>>                  "called ...\n");
>>
>>   if ( ! m_TrackingThreadStarted && this->GetThreadingEnabled() )
>>     {
>>     m_ThreadID = m_Threader->SpawnThread( TrackingThreadFunction, 
>> this );
>>     m_TrackingThreadStarted= true;
>>     }
>>
>>   m_PulseGenerator->RequestStart();
>> }
>>
>> /** Needs to be called every time when exiting tracking state. */ 
>> void Tracker::ExitTrackingStateProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::ExitTrackingStateProcessing "
>>                  "called ...\n");
>>
>>   // by default, we will exit tracking by terminating tracking thread
>>   this->ExitTrackingTerminatingTrackingThread();
>> }
>>
>> /** Exit tracking by terminating tracking thread */ void 
>> Tracker::ExitTrackingWithoutTerminatingTrackingThread( void )
>> {
>>   igstkLogMacro( DEBUG,     
>> "igstk::Tracker::ExitTrackingWithoutTerminatingTrackingThread "
>>     "called ...\n");
>>
>>   m_PulseGenerator->RequestStop();
>> }
>>
>> /** Exit tracking by terminating tracking thread */ void 
>> Tracker::ExitTrackingTerminatingTrackingThread( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::ExitTrackingTerminatingTrackingThread "
>>                  "called ...\n");
>>
>>   m_PulseGenerator->RequestStop();
>>
>>   // Terminating the TrackingThread.
>>   if ( this->GetThreadingEnabled() )
>>     {
>>     m_Threader->TerminateThread( m_ThreadID );
>>     m_TrackingThreadStarted = false;
>>     }
>> }
>>
>> /** The "AttemptToUpdateStatus" method attempts to update status
>>     during tracking. */
>> void Tracker::AttemptToUpdateStatusProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::AttemptToUpdateStatusProcessing "
>>                         "called ...\n");
>>
>>   // Set all tools to "not updated"
>>   //
>>   typedef TrackerToolsContainerType::iterator  InputConstIterator;
>>
>>   InputConstIterator inputItr = m_TrackerTools.begin();
>>   InputConstIterator inputEnd = m_TrackerTools.end();
>>
>>   unsigned int toolId = 0;
>>
>>   while( inputItr != inputEnd )
>>     {
>>     (inputItr->second)->SetUpdated(false);
>>     ++inputItr;
>>     }
>>  
>>   // wait for a new transform to be available, it would be nice if
>>   // "Wait" had a time limit like pthread_cond_timedwait() on Unix or
>>   // WaitForSingleObject() on Windows
>>   if ( this->GetThreadingEnabled() )
>>     {
>>     m_ConditionNextTransformReceived->Wait(       & 
>> m_LockForConditionNextTransformReceived );
>>     }
>>   else
>>     {
>>     this->InternalThreadedUpdateStatus();
>>     }
>>
>>   ResultType result = this->InternalUpdateStatus();
>>
>>   m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>> /** This method is called when a call to UpdateStatus succeeded */
>> void Tracker::UpdateStatusSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::UpdateStatusSuccessProcessing "
>>                  "called ...\n");
>>
>>   typedef TrackerToolsContainerType::iterator  InputConstIterator;
>>
>>   InputConstIterator inputItr = m_TrackerTools.begin();
>>   InputConstIterator inputEnd = m_TrackerTools.end();
>>
>>   while( inputItr != inputEnd )
>>     {
>>     if ( (inputItr->second)->GetUpdated() &&
>>            ( !m_ApplyingReferenceTool || 
>> m_ReferenceTool->GetUpdated() ) )       {
>>       const TransformType transform = 
>> (inputItr->second)->GetRawTransform();
>>
>>       const double timeToExpiration = transform.GetExpirationTime() - 
>>                                       transform.GetStartTime();
>>
>>       TransformType::VersorType rotationRaw;
>>       TransformType::VectorType translationRaw;
>>
>>       translationRaw = transform.GetTranslation();
>>       rotationRaw    = transform.GetRotation();
>>
>>       TransformType toolRawTransform;
>>       toolRawTransform.SetTranslationAndRotation( translationRaw, 
>> rotationRaw,
>>                         transform.GetError(),
>>                         timeToExpiration );
>>
>>       (inputItr->second)->SetRawTransform( toolRawTransform );
>>  
>>       TransformType toolCalibrationTransform
>>         = (inputItr->second)->GetCalibrationTransform();
>>      
>>       TransformType toolCalibratedTransform;
>>       toolCalibratedTransform =
>>          toolCalibratedTransform.TransformCompose(             
>> toolRawTransform, toolCalibrationTransform );
>>
>>       (inputItr->second)->SetCalibratedTransform( 
>> toolCalibratedTransform );
>>
>>       //throw an event
>>       (inputItr->second)->InvokeEvent( 
>> TrackerToolTransformUpdateEvent() );
>>               // if a reference tracker tool has been specified, then 
>> if the tracker
>>       // tool that is being updated is the selected reference tracker 
>> tool,
>>       // then update the transform that is from the tracker to the the
>>       // reference tracker tool. Otherwise, update the transform from 
>> the
>>       // tracker tool to the tracker.       if( 
>> m_ApplyingReferenceTool )
>>         {
>>         if ( (inputItr->first) == 
>> m_ReferenceTool->GetTrackerToolIdentifier())
>>           {
>>           this->RequestSetTransformAndParent(             
>> toolCalibratedTransform.GetInverse(), inputItr->second );
>>           }
>>         else
>>           {
>>           (inputItr->second)->RequestSetTransformAndParent( 
>>             toolCalibratedTransform, this );
>>           }
>>         }
>>       else
>>         {
>>         (inputItr->second)->RequestSetTransformAndParent(             
>> toolCalibratedTransform, this );
>>         }
>>       }
>>     ++inputItr;
>>     }
>>
>>   this->InvokeEvent( TrackerUpdateStatusEvent() );  }
>>
>> /** This method is called when a call to RequestUpdateStatus failed */
>> void Tracker::UpdateStatusFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::UpdateStatusFailureProcessing "
>>                  "called ...\n");
>>
>>   this->InvokeEvent( TrackerUpdateStatusErrorEvent() );  }
>>
>>
>> /** The "CloseFromTrackingStateProcessing" method closes tracker in
>>  *  use, when the tracker is in tracking state. */
>> void Tracker::CloseFromTrackingStateProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, 
>> "igstk::Tracker::CloseFromTrackingStateProcessing "
>>                  "called ...\n");
>>
>>   // leaving TrackingState, going to AttemptingToCloseState
>>   this->ExitTrackingStateProcessing();
>>
>>   ResultType result = this->InternalStopTracking();
>>
>>   // detach all the tracker tools from the tracker
>>   this->DetachAllTrackerToolsFromTracker();
>>
>>   // Terminating the TrackingThread and if it is started
>>   if ( m_TrackingThreadStarted && this->GetThreadingEnabled() )
>>     {
>>     m_Threader->TerminateThread( m_ThreadID );
>>     m_TrackingThreadStarted = false;
>>     }
>>
>>
>>   if( result == SUCCESS )
>>     {
>>     result = this->InternalClose();
>>     }
>>
>>   m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>> /** Detach all tracker tools from the tracker */
>> void Tracker::DetachAllTrackerToolsFromTracker()
>> {
>>
>>   typedef TrackerToolsContainerType::iterator  InputConstIterator;
>>
>>   InputConstIterator inputItr = m_TrackerTools.begin();    
>> InputConstIterator inputEnd = m_TrackerTools.end();
>>
>>   while( inputItr != inputEnd )
>>     {
>>     this->RemoveTrackerToolFromInternalDataContainers( 
>> inputItr->second );     ++inputItr;
>>     }
>>
>>   m_TrackerTools.clear();
>> }
>>  
>> /** The "CloseFromCommunicatingStateProcessing" method closes
>>  *  tracker in use, when the tracker is in communicating state. */
>> void Tracker::CloseFromCommunicatingStateProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::"
>>                  "CloseFromCommunicatingStateProcessing called ...\n");
>>
>>   // Detach all the tracker tools from the tracker
>>   this->DetachAllTrackerToolsFromTracker();
>>
>>   // Terminating the TrackingThread and if it is started
>>   if ( m_TrackingThreadStarted && this->GetThreadingEnabled() )
>>     {
>>     m_Threader->TerminateThread( m_ThreadID );
>>     m_TrackingThreadStarted = false;
>>     }
>>
>>     ResultType result = this->InternalClose();
>>
>>   m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>>
>> /** Post-processing after close tracking has been successful. */ void 
>> Tracker::CloseCommunicationSuccessProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::"
>>                  "CloseCommunicationSuccessProcessing called ...\n");
>>
>>   this->InvokeEvent( TrackerCloseEvent() );
>> }
>>
>> /** Post-processing after close tracking has failed. */ void 
>> Tracker::CloseCommunicationFailureProcessing( void )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::"
>>                  "CloseCommunicationFailureProcessing called ...\n");
>>
>>   this->InvokeEvent( TrackerCloseErrorEvent() );
>> }
>>
>> /** Print object information */
>> void Tracker::PrintSelf( std::ostream& os, itk::Indent indent ) const
>> {
>>   Superclass::PrintSelf(os, indent);
>>
>>   if( this->m_PulseGenerator )
>>     {
>>     os << indent << this->m_PulseGenerator << std::endl;
>>     }
>>
>>   if( this->m_PulseObserver )
>>     {
>>     os << indent << this->m_PulseObserver << std::endl;
>>     }
>>
>>   os << indent << "ValidityTime: " << this->m_ValidityTime << std::endl;
>>   os << indent << "CoordinateSystemDelegator: ";
>>   this->m_CoordinateSystemDelegator->PrintSelf( os, indent );
>> }
>>
>> /** The "SetFrequencyProcessing" passes the frequency value to the Pulse
>>  * Generator. Note that it is still possible for the PulseGenerator 
>> to reject
>>  * the value and stay at its current frequency. */
>> void Tracker::SetFrequencyProcessing( void )
>> {
>>   igstkLogMacro( DEBUG,                  
>> "igstk::Tracker::SetFrequencyProcessing called ...\n");
>>
>>   this->m_PulseGenerator->RequestSetFrequency( 
>> this->m_FrequencyToBeSet );
>> }
>>
>>
>> /** Request adding a tool to the tracker  */
>> void
>> Tracker::
>> RequestAttachTool( TrackerToolType * trackerTool )
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::"
>>                  "RequestAttachTool called ...\n");
>>  
>>   m_TrackerToolToBeAttached = trackerTool;
>>     igstkPushInputMacro( AttachTrackerTool );
>>   this->m_StateMachine.ProcessInputs();
>> }
>>
>> /** The "AttemptToAttachTrackerToolProcessing" method attempts to
>>  * add a tracker tool to the tracker */
>> void Tracker::AttemptToAttachTrackerToolProcessing( void )
>> {
>>   igstkLogMacro( DEBUG,     
>> "igstk::Tracker::AttemptToAttachTrackerToolProcessing called ...\n");
>>
>>   // Verify the tracker tool information before adding it to the
>>   // tracker. The conditions that need be verified depend on   // the 
>> tracker type. For example, for MicronTracker, the   // the tracker 
>> should have access to the template file of the
>>   // Marker that is attached to the tracker tool.   ResultType result 
>> =     this->VerifyTrackerToolInformation( m_TrackerToolToBeAttached );
>>  
>>   m_StateMachine.PushInputBoolean( (bool)result,
>>                                    m_SuccessInput,
>>                                    m_FailureInput );
>> }
>>
>> /** Request remove a tool from the tracker  */
>> Tracker::ResultType Tracker::
>> RequestRemoveTool( TrackerToolType * trackerTool )
>> {
>>   this->m_TrackerTools.erase( trackerTool->GetTrackerToolIdentifier() );
>>   this->RemoveTrackerToolFromInternalDataContainers( trackerTool );   
>> return SUCCESS;
>> }
>>
>> const Tracker::TrackerToolsContainerType &
>> Tracker::GetTrackerToolContainer() const
>> {
>>   return m_TrackerTools;
>> }
>>
>> /** Thread function for tracking */
>> ITK_THREAD_RETURN_TYPE Tracker::TrackingThreadFunction(void* 
>> pInfoStruct)
>> {
>>   struct itk::MultiThreader::ThreadInfoStruct * pInfo =     (struct 
>> itk::MultiThreader::ThreadInfoStruct*)pInfoStruct;
>>
>>   if( pInfo == NULL )
>>     {
>>     return ITK_THREAD_RETURN_VALUE;
>>     }
>>
>>   if( pInfo->UserData == NULL )
>>     {
>>     return ITK_THREAD_RETURN_VALUE;
>>     }
>>
>>   Tracker *pTracker = (Tracker*)pInfo->UserData;
>>
>>   // counters for error rates
>>   unsigned long errorCount = 0;
>>   unsigned long totalCount = 0;
>>
>>   int activeFlag = 1;
>>   while ( activeFlag )
>>     {
>>     ResultType result = pTracker->InternalThreadedUpdateStatus();
>>     pTracker->m_ConditionNextTransformReceived->Signal();
>>         totalCount++;
>>     if (result != SUCCESS)
>>       {
>>       errorCount++;
>>       }
>>           // check to see if we are being told to quit     
>> pInfo->ActiveFlagLock->Lock();
>>     activeFlag = *pInfo->ActiveFlag;
>>     pInfo->ActiveFlagLock->Unlock();
>>     }
>>     igstkLogMacroStatic(pTracker, DEBUG, "TrackingThreadFunction was "
>>                       "terminated, " << errorCount << " errors "
>>                       "out of " << totalCount << "updates." << 
>> std::endl );
>>
>>   return ITK_THREAD_RETURN_VALUE;
>> }
>>
>> /** Report to the tracker tool that the tool is not available */
>> void Tracker::ReportTrackingToolNotAvailable( TrackerToolType * 
>> trackerTool ) const
>> {
>>   igstkLogMacro( DEBUG,     
>> "igstk::Tracker::ReportTrackingToolNotAvailable called...\n");
>>   trackerTool->RequestReportTrackingToolNotAvailable();
>> }
>>
>> /** Report to the tracker tool that the tool is Visible */
>> void Tracker::ReportTrackingToolVisible( TrackerToolType * 
>> trackerTool ) const
>> {
>>   igstkLogMacro( DEBUG, "igstk::Tracker::ReportTrackingToolVisible 
>> called...\n");
>>   trackerTool->RequestReportTrackingToolVisible();
>> }
>>
>> /** Set raw transform */
>> void Tracker::SetTrackerToolRawTransform(   TrackerToolType * 
>> trackerTool, const TransformType transform )
>> {
>>   igstkLogMacro( DEBUG,     
>> "igstk::Tracker::SetTrackerToolRawTransform called...\n");
>>   trackerTool->SetRawTransform( transform );
>> }
>>
>> /** Turn on/off update flag of the tracker tool */
>> void Tracker::SetTrackerToolTransformUpdate(   TrackerToolType * 
>> trackerTool, bool flag ) const
>> {
>>   igstkLogMacro( DEBUG,      
>> "igstk::Tracker::SetTrackerToolTransformUpdate called...\n");
>>   trackerTool->SetUpdated( flag ); }
>>
>> /** Report invalid request */
>> void Tracker::ReportInvalidRequestProcessing( void )
>> {
>>   igstkLogMacro( DEBUG,     
>> "igstk::Tracker::ReportInvalidRequestProcessing called...\n");
>>
>>   this->InvokeEvent( InvalidRequestErrorEvent() );
>> }
>>
>> }
>>
>>
>> ------------------------------------------------------------------------
>>
>> _______________________________________________
>> IGSTK-Users mailing list
>> IGSTK-Users at public.kitware.com
>> http://public.kitware.com/cgi-bin/mailman/listinfo/igstk-users

-- 
Vincent Gratsac
----------------------------------------------------
Unité/Projet VisAGeS U746
INSERM/INRIA/CNRS/U. de Rennes I
IRISA
Campus de Beaulieu, 35042 Rennes Cedex, France

Ph:  +33 (0) 2 23 23 49 20/ Fax: +33/0 2 99 84 71 71
email: Vincent.Gratsac at irisa.fr
http://www.irisa.fr/visages/
-----------------------------------------------------




More information about the IGSTK-Users mailing list