[IGSTK-Users] Viewer Problem due to TransformPhysicalPointToIndex
Patrick Cheng
cheng at isis.georgetown.edu
Wed Jun 14 10:56:26 EDT 2006
Hi Fucang,
You are right, we are using the ITK image's
TransformPhysicalPointToIndex function, which does the truncation.
I will try to change it to TransformPhysicalPointToContinuousIndex, see
if it will do any better.
Thank you for the detailed email.
Patrick
Fucang Jia wrote:
> Hi, Patrick,
>
> There is a problem due to TransformPhysicalPointToIndex function, which
> belongs to Bug# 738,
> but it seems that this critial bug has not been fixed.
>
> When I load IGSTK\Sandbox-Release-8\Testing\Data\E000192,
> press the left mouse button on the sagittal view in
> FourViewsTrackingWithCT app continuously,
> then the index changes little bye one, when the index of sagittal viewer
> was changed to 0, I press the
> mouse button onto the sagittal image, it says "Picked point outside
> image". I tracked this bug,
> and found that it is due to index round up problem.
>
> For example, the origin is -153.6640625,
> the spacing is 0.671875
> the physical coordinate is 16.3203
> so the according index is
> (16.3203+153.6640625)/0.671875 = 252.9999813953,
> it was truncated to 252.
>
> Then resliceImage function will change the orthogonal image slice.
>
> I also found this behaviour will occur when certain DICOM dataset is
> loaded. When I load other DICOM dataset,
> the axial and coronal slice will have problem, but sagittal view does
> not change when mouse pressed. It is because
> that certain origin and spacing parameters trigger this problem.
>
> Thanks!
>
> Fucang
>
>
> ------------------------------------------------------------------------
>
> _______________________________________________
> IGSTK-Users mailing list
> IGSTK-Users at public.kitware.com
> http://public.kitware.com/cgi-bin/mailman/listinfo/igstk-users